Let me know if you can extend many meters of cables of the current sensors.
The current sensors can be installed in junction boxes, together with other cables must be isolated
length current sensors
Submitted by Guest on Sun, 18/11/2012 - 10:33»
Re: length current sensors
I think you should keep them as short a possible. Can someone confirm if this is true?
Re: length current sensors
Some considerations:
If the CT's are located too far away from the emonTx then you should consider loss in the current measured and interference from other electromagnetic sources like other high current conductors in the vicinity.
Depending on the distance you need you can try to use a very good quality shielded wires, like coaxial cables.
Since the impedance presented by the emonTx is very low (the burden resistor mainly, which is around 20 to 50 ohms) the type of cable wont`t make substantial differences, just use a good one but keep them as short as possible anyway.
Hope this helps.
Re: length current sensors
Two points.
The main problem is likely to be interference from other cables and switches. Avoid running the cable parallel to any other cable unless at least 1 m separates the two cables, if it must cross, cross at 90°.
A cable like this http://uk.rs-online.com/web/p/multicore-microphone-instrument-cable/2074... should be suitable.
You should keep the cable as short as possible.
Re: length current sensors
so if there are other cables near to interference?
how far you have to find??